Vintage465 wrote:
I checked the amp draw on both of the units. The amp draw must include the Humidifier. If it didn't the two units would use 106.72 amps over an 8hr period. That would have taken my to 50% of their capacity. That has not happened. I have used the two Phillips Respironic units quite a few times. The Resmed unit uses about 1.25 amp hr. more. I am thinking with my Phillips Respironics and the Resmed unit both running w/o Humidifiers on would be just about the same(it will draw 10a more over night) as the two Phillips units w/o the Humidifiers on. Thinking it will be OK.
Finishing up from my perspective..........the one thing I am not good at is unplugging the unit from the 12v for traveling. I have logged quite a few miles with it plugged in, but I will need to get better. These things are a bit too pricey to be irresponsible with!
and with the humidifier, the current draw is NOT a constant function of time. It is high initally to heat the water, then will drop way down and the heating unit will cycle to keep the water at a constant temperature. How much cycling and how much current will depend on air flow rate, and air temperature as well as outside temperature and how much water remains in the reservoir. as water level drops energy needed to keep it at a constant temperature decreases.
But the spec's are for worst case max draw conditions, likely full reservoir, highest airflow rate, lowest outside temperature, very low humidity and line voltage that will draw max current. Then a safety factor added to that.
Having designed test equipment to meet UL spec's one NEVER NEVER EVER wants to ship a unit that draws more current than the nameplate spec. That will get you in big trouble with U.L.
Probably even worse for medical equipment.
So. I'd take the nameplate current draw as a serious absolute worst case operating condition.
I've found the easiest way to estimate the actual AH draw is to read my trimetric before going to bed and in the AM to see battery SOC and then calculate AH useage. Need to do that in the evening when everything else is off and in the AM before turning on the heat etc.
And I've run mine off a 200WH Lithium Ion battery pack and seen how many hours till it quites. Usually sometime in the third night. That works out to between 1 and 1.5A, consistent with what I get from the trimetric.
It's not peak Amps you care about it's the Amps integrated over the time it is on.